The new MIKE_blue CCD currently begins to go
non-linear at a pixel count of approximately 8500 DN and is strongly
non-linear by a pixel count of 10000 DN. (This is measured for single
pixels .... i.e. binning of 1x1) The limitation appears to be in the
pixel full well and not in the electronics, therefore, higher counts
can be obtained -up to the digital saturation - by binning.
The folllowing table provides measured gain and read noise values at
various count levels for 1x1 binned flat field observations using the
internal incandescent lamp. The mean pixel count is measured in a
region of about 30x70 pixels and has not had the bias level of about
420 DN subtracted.
Pixel_Count(mean[DN]) Gain(e-/DN) Noise(e-)
1592 0.46 1.68
2715 0.46 1.67
2898 0.45 1.62
3896 0.43 1.58
4306 0.47 1.74
4610 0.42 1.56
4850 0.46 1.67
5308 0.46 1.66
5960 0.44 1.61
7280 0.47 1.72
8086 0.51 1.89
8667 0.54 1.97
9080 0.64 2.33
9849 1.39 5.08
10300 3.28 12.06
11049 9.13 33.4
These values are taken from measurements across several flat field
pairs of varying integration times. The pixel threshold value is
approximately 11000 DN.
Test flats with an exposure time of 30 seconds have peak pixel values
of about 5000 counts and have gain and noise measuremnets across the
orders ranging from 0.44-0.50 (1.57-1.88) at count levels of 250-4500.
Bottom Line: MIKE_Blue is effective only for fainter targets. The
current configuation may not be able to be used on bright target
observations (once the single pixel count levels are above about 4000
electrons.
MIKE_Blue limited dynamic range
|
|